AKTA purifier Prgramming - (Nov/21/2008 )

Hello Everyone
I am new user to AKTA purifier and wish to programme it as follow-
start concentration of B 45% should run for 100 ml volume
then Concentration of B should reach to 60% and run for 50 ml
then concentration of B should reach to 100% in next 50 ml (where protein will be eluted from column)

Can anybody please suggest me which type of gradient i should opt for on AKTA
I tried to use segmented gradient but machine got stucked at 45% of B and it never went above that concentration. I tried several times reprogramming it in vain.

It will great if somebody please give suggestions in this context

Thanks
rang

-rang-

did you program multiple gradient steps?

which version of unicorn are you using (i have an old version).

-mdfenko-

I can't remember if the gradient setting up according to the volume or the time.

I think it is a time issue. I will check it tomorrow when I go to work.

So set 45% B 100ml ( for example, if your flow rate is 2ml/min, then you should do 45% B, 50min), etc...

-Qiuhong-