ChIP sonication problem: too wide range of smear - too wide range of smear (Jun/05/2006 )
Hi All,
I am just starting to do ChIP. My problem with the sonication is that when I tested the sonicated DNA on 1% gel, I saw a smear ranging from 200 bp to 4kb (even after I treated with RNase A). There were no bands, though, just smear. Any suggestion will be appreciated. Thanks!!!
--Sure
I am just starting to do ChIP. My problem with the sonication is that when I tested the sonicated DNA on 1% gel, I saw a smear ranging from 200 bp to 4kb (even after I treated with RNase A). There were no bands, though, just smear. Any suggestion will be appreciated. Thanks!!!
--Sure
I don't think the smear is RNA, but simly genomic DNA that you have successfully sheared...
Thanks!
I don't think the RNA is a problem, either. However, my question is, how to make the smear range less wide, like within 500bp-1KB. This is important for my experiment, because I will use the ChIP products on microarray.
Thanks again!!
I don't think the RNA is a problem, either. However, my question is, how to make the smear range less wide, like within 500bp-1KB. This is important for my experiment, because I will use the ChIP products on microarray.
Thanks again!!
Hello,
To get DNA in the 1kb-500 bp range I used to sonicate 5 times 15s at 5 level Power on a XL Misonix. Between each sonication I put my samples back into ice.
If it is not enough you can maybe try to increase the number of sonication.
Hope it helps.
Vero
hi sure,
not too sure which sonication machine you are using, but with a branson sonifier with a microtip attachment, at 30% duty cycle, 10 15 second pulses usually does the trick for me, i work with mammalian cells. a lower setting on your sonication machine may do the trick.
Nick
not too sure which sonication machine you are using, but with a branson sonifier with a microtip attachment, at 30% duty cycle, 10 15 second pulses usually does the trick for me, i work with mammalian cells. a lower setting on your sonication machine may do the trick.
Nick
Hello,
I am going to be starting ChIP analysis soon, could someone with more experience provide a gel image shot of what good shearing looks like. I am unsure how discrete a band I should see around 500bp.
Thanks
Hello,
I am going to be starting ChIP analysis soon, could someone with more experience provide a gel image shot of what good shearing looks like. I am unsure how discrete a band I should see around 500bp.
Thanks
here it is http://www.protocol-online.org/forums/inde...showtopic=12700
I think the wide range of smear is indicative of inadequate sonication. If you sonicate more, you will see a smear from 1 kb to 200 bp with 500 bp being the most intense area.
Thanks very much for the prompt reply PCRman. So lane 4 would be a good sonication.
A potential cause is that the volume of your samples is too big so that the DNA is not evenly sonicated.