Method for protein concentration measurement - (Sep/06/2012 )
Hi, everyone!
I purified a protein several days ago and measured the concentration of the sample yesterday using two different method. First, it was measured by BCA kit. the concentration is 7.6mg/ml. Then I measured UV280 of the sample and the concentration was calculated to be 14.3mg/ml(I know the Ext. coefficient). This is very strange! I don't know which method is more convincible. Which method do you usually use for protein concentration measurement?
I am looking forward to your replies. Thank you!
factors that could affect your protein determination:
standard protein used to calibrate the bca
buffer in which the protein is solubilized
absorbance reading (linear range of the assay, dynamic range of the spectrophotometer)
dilution of the sample (to stay within effective range)
purity of the protein
My money would be on the BCA reading. It would be unusual for BCA to give a dramatically low reading unless you have a copper chelator present.
If you know the amino acid sequence then I would trust the Native scan. You could determine the concentration by denaturing your protein with some NaOH and using the extinction coefficient of your protein with the Tyrosines and Tryptophans at 260nm (if you have these amino acids).
Smog187 on Fri Sep 14 02:32:08 2012 said:
If you know the amino acid sequence then I would trust the Native scan. You could determine the concentration by denaturing your protein with some NaOH and using the extinction coefficient of your protein with the Tyrosines and Tryptophans at 260nm (if you have these amino acids).
mdfenko on Fri Sep 14 19:26:25 2012 said:
Smog187 on Fri Sep 14 02:32:08 2012 said:
If you know the amino acid sequence then I would trust the Native scan. You could determine the concentration by denaturing your protein with some NaOH and using the extinction coefficient of your protein with the Tyrosines and Tryptophans at 260nm (if you have these amino acids).
Smog187 on Fri Sep 14 02:32:08 2012 said:
If you know the amino acid sequence then I would trust the Native scan. You could determine the concentration by denaturing your protein with some NaOH and using the extinction coefficient of your protein with the Tyrosines and Tryptophans at 260nm (if you have these amino acids).
Thank you very much! Could you give some suggestions about how much NaOH is needed?
Rchqing on Sat Sep 15 05:39:03 2012 said:
Smog187 on Fri Sep 14 02:32:08 2012 said:
If you know the amino acid sequence then I would trust the Native scan. You could determine the concentration by denaturing your protein with some NaOH and using the extinction coefficient of your protein with the Tyrosines and Tryptophans at 260nm (if you have these amino acids).
Thank you very much! Could you give some suggestions about how much NaOH is needed?
Smog187 on Mon Sep 17 00:32:51 2012 said:
Rchqing on Sat Sep 15 05:39:03 2012 said:
Smog187 on Fri Sep 14 02:32:08 2012 said:
If you know the amino acid sequence then I would trust the Native scan. You could determine the concentration by denaturing your protein with some NaOH and using the extinction coefficient of your protein with the Tyrosines and Tryptophans at 260nm (if you have these amino acids).
Thank you very much! Could you give some suggestions about how much NaOH is needed?
OK! Thank you very much!