Protocol Online logo
Top : New Forum Archives (2009-): : SDS-PAGE and Western Blotting

adjusting pH of 4x stacking buffer - (Aug/14/2012 )

I am following a protocol that requires the stacking buffer be adjusted to a pH of 6.8. I accidentally overadjusted it to around 6.1 so I used some NaOH to compensate the pH back to 6.8. Would this affect the stacking gel performance or not really?

-pinkc-

The NaOH adds Na to your solution (the OH- combines with the H+ in the solution to form water), if the assay is dependent on salt you may have a problem, but it should be fine most of the time.

The usual way to do this sort of thing is to make a tris solution of pH 6.8 and add that to your stacking buffer.

-bob1-

the addition of naoh will affect the stacking buffer.

if you overshoot then you should add more basic component (tris), calculated for a greater volume, readjust pH, then bring to the new final volume.

-mdfenko-

Ack! OK thanks a lot for replying - I haven't used it yet but now I will dump it out and remake it and be careful about adjusting pH.

-pinkc-