problem of membrane protein FPLC - (Feb/28/2010 )

I purified my membrane protein by FPLC gel filtration column. Before the injection , I checked the OD280, which is around 1AU. then i took 0.5ml protein sample to run the FPLC. However, no sharp peak came out. I just got a broad and small peak with 35 mAu. which is much lower than that expected. (I purified the membrane protein with DM and the FPLC buffer : 150mM NaCl, 20mM HEPEs, 0.4g/100ml DM)

Is it that my protein has aggregated on the FPLC column?

I also try to increase salt concentration (300mM) of the FPLC buffer , then I got a sharp peak with OD 160 mAu and the retention time is some early than Void Volume. Is it normal ? anybody can give me some suggestion for the FPLC pufication condition?

-handsome-

which fplc column did you use?

recommended minimum ionic strength for fplc buffer (gel filtration) is 200mM. this prevents non-specific binding to the stationary phase.

if the peak came before the void then it was something released from the column prior to the injection. when you re-equilibrated with the higher salt buffer you probably did not completely wash the column before you injected (anything bound that was released will fractionate).

-mdfenko-