Hi,

Has anyone tried using this kit from Amersham (ECL Western bolotting detection reagent RPN2209)? Can the image on the membrane be captured using light box instead of on X-ray films?

Thanks.

-SilverSta-

If by light box you mean something like the Epichemi II darkroom (UVP Lab Products, UK) which has a CCD camera with appropriate filters or Biorads VersaDoc system then the answer is yes. I actually found the Epichem more sensitive that film

-DevGrp-

Thanks DevGrp. Unfortunately my lab doesn't have any of those. :D

-SilverSta-

Well in therory any sensitive CCD camera with an appropriate filter and in some kind of dark room / dark box should work.
The only problem would be getting software to allow you to collect the image over several minutes (5-15min)

-DevGrp-

As an aside, if you want to save your lab lots of money, then you can make your own ECL reagents.

they work just as well as ECL reagents but at a fraction of the cost.

Solution 1

1ml 250mM luminol
0.44ml 90mM p-coumaric acid
10ml 1M tris pH 8.5
up to 100ml with milliQ water

Solution 2
64ul 30% hydrogen peroxide
10ml 1M tris pH8.5
up to 100ml with milliQ water

you use an equal volume of each solution on a blot, typically 1ml of each and leave on for 1 minute.
As the above solutions don't have preservatives, they won't last for months and months but I've used them up to 2months after preparing and they still worked fine