Hi there,
I agree with all that is mentioned above regarding the speed, 5000rpm is probably too much, however one more thing you could check is that all this does not have anything to do with your 1.5ml centrifuge tubes.
Is anyone else in your lab having similar issues?
At some point we had this problem in our lab and were very confused with not getting the expected pellets.
We had recently changed the supplier we had for the tubes. We connected the dots finally and tried harvesting cells from 2 duplicate wells and pelleting them using tubes from our previous and new supplier.
There was a nicely formed pellet in the "good" tubes and almost nothing visible in the "bad" tubes. There were cells smeared on the wall of the tube, but that's not good enough really.
If in your case it has anything to do with the tubes, I would suggest -if you cannot change the tubes obviously!- to spin at lower rpm and for longer, like 6-8mins, which seemed to help a bit.
Good luck!
ioannap