measuring nucleic acid concentration on nanodrop


  • Please log in to reply
8 replies to this topic

#1 starvingstudent

starvingstudent

    member

  • Active Members
  • Pip
  • 5 posts
0
Neutral

Posted 21 July 2009 - 07:46 PM

I have a question regarding measuring concentration on the nanodrop. Does it matter how much volume (more than 1 microliter) I load onto the machine? If I load more than 1 microliter onto the nanodrop,say 2 microliters of a DNA sample and the resulting concentration is 140 ng/microliter, does that mean that there are 70 ng in each microliter of my sample or is the concentration the machine gives me (140 ng/microliter) the actual concentration of my sample regardless of how much I load?
Thanks in advance for any clarification you all can provide!

#2 Vini

Vini

    Veteran

  • Active Members
  • PipPipPipPipPipPipPipPipPipPip
  • 178 posts
1
Neutral

Posted 21 July 2009 - 09:56 PM

I have a question regarding measuring concentration on the nanodrop. Does it matter how much volume (more than 1 microliter) I load onto the machine? If I load more than 1 microliter onto the nanodrop,say 2 microliters of a DNA sample and the resulting concentration is 140 ng/microliter, does that mean that there are 70 ng in each microliter of my sample or is the concentration the machine gives me (140 ng/microliter) the actual concentration of my sample regardless of how much I load?
Thanks in advance for any clarification you all can provide!


No, it does not matter whether u load 1 or 2 ul. Thats because u get the concentration in ng/ul. therefore, volume is automatically taken care of. to remove pipetting error, i prefer taking 2 ul of my sample.

#3 jiajia1987

jiajia1987

    Veteran

  • Active Members
  • PipPipPipPipPipPipPipPipPipPip
  • 212 posts
0
Neutral

Posted 21 July 2009 - 10:12 PM

THe concentration that nanodrop gives is in per ul, so it doesnt matter how much you put in. I use 2uL too.

#4 swanny

swanny

    Veteran

  • Active Members
  • PipPipPipPipPipPipPipPipPipPip
  • 367 posts
8
Neutral

Posted 22 July 2009 - 05:50 PM

Ditto, 2 ul. 1 ul is a bit fiddly.
Heart disease kills more women than breast cancer, but heart attack symptoms differ from men's symptoms. Get to know your heart... it could save your life.

#5 fishdoc

fishdoc

    Veteran

  • Active Members
  • PipPipPipPipPipPipPipPipPipPip
  • 272 posts
2
Neutral

Posted 22 July 2009 - 05:52 PM

Agreed. 2 ul for the sample and 2 ul for the water blank. Only use 1.5 ul to initialize the machine.

#6 jiajia1987

jiajia1987

    Veteran

  • Active Members
  • PipPipPipPipPipPipPipPipPipPip
  • 212 posts
0
Neutral

Posted 22 July 2009 - 06:16 PM

Agreed. 2 ul for the sample and 2 ul for the water blank. Only use 1.5 ul to initialize the machine.


Hi fishdoc,

What I always do is to add 2ul and den close the cover and initialize the machine. After that is done, I just click on blank. This is alright, right? That was what my supervisor told me. Didnt know that I had to take a small volume to initialize first, den clean the machine, and add again to blank.

#7 fishdoc

fishdoc

    Veteran

  • Active Members
  • PipPipPipPipPipPipPipPipPipPip
  • 272 posts
2
Neutral

Posted 22 July 2009 - 06:29 PM

Agreed. 2 ul for the sample and 2 ul for the water blank. Only use 1.5 ul to initialize the machine.


Hi fishdoc,

What I always do is to add 2ul and den close the cover and initialize the machine. After that is done, I just click on blank. This is alright, right? That was what my supervisor told me. Didnt know that I had to take a small volume to initialize first, den clean the machine, and add again to blank.




I'm sure that's fine as well. We have a central facility that has a bunch of equipment, including the nanodrop, and that's what their protocol said to do (how I do it). I don't see a reason that what you're doing would be a problem. However, like I said, I just use the machine per their protocol, I've never sat down and read the company's instructions for use.

#8 jiajia1987

jiajia1987

    Veteran

  • Active Members
  • PipPipPipPipPipPipPipPipPipPip
  • 212 posts
0
Neutral

Posted 26 July 2009 - 05:52 PM

Agreed. 2 ul for the sample and 2 ul for the water blank. Only use 1.5 ul to initialize the machine.


Hi fishdoc,

What I always do is to add 2ul and den close the cover and initialize the machine. After that is done, I just click on blank. This is alright, right? That was what my supervisor told me. Didnt know that I had to take a small volume to initialize first, den clean the machine, and add again to blank.




I'm sure that's fine as well. We have a central facility that has a bunch of equipment, including the nanodrop, and that's what their protocol said to do (how I do it). I don't see a reason that what you're doing would be a problem. However, like I said, I just use the machine per their protocol, I've never sat down and read the company's instructions for use.


okay thanks~

#9 microgirl

microgirl

    Veteran

  • Active Members
  • PipPipPipPipPipPipPipPipPipPip
  • 122 posts
0
Neutral

Posted 31 July 2009 - 10:09 AM

You should be blanking your machine with whatever your nucleic acid is in - ie initialize with the water then clean that off and blank it with EB buffer, TE buffer, or whatever your n.a. is in. Only blank on water if it's suspended in water.