I using EP get the siRNA into the cells, it works by checking the protein level. However I still could get the mRNA level by QPCR.
the Ct value is simalr to control.
What is wrong with the QPCR or siRNA or my primer?
why using siRNA knock down,but get the mRNA products?
Started by ucsd, Jul 06 2009 11:54 PM
3 replies to this topic
#1
Posted 06 July 2009 - 11:54 PM
#2
Posted 07 July 2009 - 09:27 AM
ucsd, on Jul 6 2009, 11:54 PM, said:
I using EP get the siRNA into the cells, it works by checking the protein level. However I still could get the mRNA level by QPCR.
the Ct value is simalr to control.
What is wrong with the QPCR or siRNA or my primer?
the Ct value is simalr to control.
What is wrong with the QPCR or siRNA or my primer?
Although siRNA may be blocking the protein synthesis, primarily siRNA should degrade the mRNA. The only reasons I can think of your qPCR not working despite mRNA degradation are..
1. You are using wrong housekeeping gene for normalization.
2. You are using primers that bind to an undegraded fragment of mRNA.
These are just thoughts, I have no real knowledge of this. If you don't get any good feedback anywhere, I suggest desiging a qPCR primer pair in another region of mRNA, and/or trying out various housekeeping genes. I would not throw away the cells just yet, since you get knockdown at a protein level.
Let us know if you solve this puzzle.
#3
Posted 07 July 2009 - 01:11 PM
I've not heard of EP, but if I understand your problem correctly, you introduce siRNA into your cells and see a reduction at the protein level, but not the mRNA level. If this is the case, then this sounds very much like a miRNA-like mechanism of translational inhibition, not degradation. Check your siRNA sequence, specifically to make sure nucleotides 9-12 base pair perfectly. If there is a mismatch/bulge then miRNA mechanism is very likely.
Of course, the suggestions put forward be cell counter are also likely/possible.
Did you design the siRNA, or pre-designed by a company? Sometimes they have a qRT-PCR assay specifically for that siRNA.
Of course, the suggestions put forward be cell counter are also likely/possible.
Did you design the siRNA, or pre-designed by a company? Sometimes they have a qRT-PCR assay specifically for that siRNA.
#4
Posted 07 July 2009 - 03:44 PM
My suggestion is design a new set of primers for qPCR or qRT-PCR. These two oligos better probe on two different exons of your target mRNA, then you might be fine.