Hi all,

Hope someone can help me with my weird week.

I tried running DNA thru 1% agarose gel and for some unknown reason they wont run. I tried many times and I keep getting weirder results.
1st time I tried both loading dye and (I assume because I checked with the UV) DNA did not run.
2nd time I tried the loading dye ran but no DNA was seen under UV

Both times were different newly prepared gel, so I am not at a total lost.

MAD :P

#2 jiajia1987

jiajia1987

    Veteran

  • Active Members
  • PipPipPipPipPipPipPipPipPipPip
  • 212 posts
0
Neutral

Posted 29 June 2009 - 06:51 PM

Did you check the composition of the gel? Is it really 1% or did you make it a higher percentage by mistake?

Did you cover the gel with enough buffer? Try not to overload it with buffer though. What was the voltage and time?

#3 mad

mad

    member

  • Active Members
  • Pip
  • 11 posts
0
Neutral

Posted 29 June 2009 - 07:41 PM

both gels were freshly made and I've done it tons of times.
I was testing the gel run rate at 45v for an hour (Don't ask me why my insane supervisor is the one behind this idea), but i've already previously done it before with success.

Oh I forgot to mention that I was running it in a smaller tank from the usual equipment (Soz but the IP crap is making it hard for me to explain my problem)

MAD :P

#4 HomeBrew

HomeBrew

    Veteran

  • Global Moderators
  • PipPipPipPipPipPipPipPipPipPip
  • 930 posts
16
Good

Posted 29 June 2009 - 08:24 PM

I presume these are EtBr stained gels? I further presume that you're looking at them with the right wavelength UV, and that the power supply was set for constant voltage, and the leads are set correctly (all mistakes I have made over the years)?

Is the power supply working? Do bubbles form along the negative wire in the chamber?

Is the sample DNA still visible in or near the wells?

Did you include a ladder, and did it run and was it visible?

#5 mad

mad

    member

  • Active Members
  • Pip
  • 11 posts
0
Neutral

Posted 29 June 2009 - 08:33 PM

How do you check if the DNA is still in or near the well?

I've looked under the UV and can see a slight brightness at the well. Is that it?

#6 gebirgsziege

gebirgsziege

    I like fungi

  • Active Members
  • PipPipPipPipPipPipPipPipPipPip
  • 392 posts
25
Excellent

Posted 29 June 2009 - 10:25 PM

What kind of DNA are you working with (i.e. plasmid, genomic, PCR-product)?
I agree with HomeBrew: what about your ladder? If the ladder is running correctly, maybe the answer is simple: your PCR did not work?

Usually you can approximate where your DNA is by spotting the dyes...e.g. Bromphenolblue is at app. 300 - 200 bp.

If using EtBr stained gels, you will see the DNA bright only if it ran into the gel. Maybe you should try to add a drop or EtBr to your sample and look at it without a gel in UV light....if there is DNA in it it should be brighter than EtBr without DNA. So you can find out if there is DNA in your sample......
A man cannot be too careful in the choice of his enemies. (Oscar Wilde)

#7 A_Gisela

A_Gisela

    member

  • Members
  • Pip
  • 4 posts
0
Neutral

Posted 02 July 2009 - 01:56 PM

Are you sure your sample is DNA?, could be that you have problems on the isolation method and you are not getting DNA.

Did you check the electrodes?, is your chamber working correctly?

#8 swanny

swanny

    Veteran

  • Active Members
  • PipPipPipPipPipPipPipPipPipPip
  • 367 posts
8
Neutral

Posted 02 July 2009 - 03:24 PM

Chuck out your running buffer, make up a fresh solution, prepare a new gel using that buffer and try again.
Heart disease kills more women than breast cancer, but heart attack symptoms differ from men's symptoms. Get to know your heart... it could save your life.

#9 mad

mad

    member

  • Active Members
  • Pip
  • 11 posts
0
Neutral

Posted 02 July 2009 - 11:11 PM

thanks for all the suggestion

I have another question to ask regarding the issue. Does the thickness of the loading dye can have an effect on the sample run?
This is because in this instance I am only getting small amount of product so I cant exactly do a 1:5 (Loading dye:Sample)dilution, so I'm currently doing a 1:1 or 1:2...depending on amount available.

Could this be the problem?


MAD :(

#10 warsel

warsel

    Enthusiast

  • Active Members
  • PipPipPipPipPip
  • 77 posts
0
Neutral

Posted 03 July 2009 - 12:34 AM

thanks for all the suggestion

I have another question to ask regarding the issue. Does the thickness of the loading dye can have an effect on the sample run?
This is because in this instance I am only getting small amount of product so I cant exactly do a 1:5 (Loading dye:Sample)dilution, so I'm currently doing a 1:1 or 1:2...depending on amount available.

Could this be the problem?


MAD :(


possible but not likely. you can also dilute the loading dye to make it 2x instead of 6x of course to counter that problem.

is your maker visible on the gel?
- if it is: 1) did you check to see if your DNA actually stays in the well; in my experience that is usually a problem with samples that have a lot of residual ethanol in them - try to add sucrose to the loading dye to counter the problem.
2) are you sure there is actually (sufficient) DNA in your tube? if yes - can you exclude that it's just free nucleotides (ie a failed PCR run)?

- if it isn't: check your buffers, agarose, etbr and elpho unit I'd say.