Centrifugation of epithelial cells


2 replies to this topic

#1 floyd78

    member

  • Active Members
  • Pip
  • 7 posts

Posted 17 June 2009 - 03:02 AM

Can fixed angle centrifuge rotors be used to spin down cultured epithelial cells? If so are any changes in normal speeds/times necessary? I currently use 220g for 4 mins in a swinging bucket system but will shortly be moving labs so will only have access to a fixed angle system temporarily.

Thanks

#2 bob1

    Hmmm, I think it's working

  • Active Members
  • PipPipPipPipPip
  • 2,740 posts

Posted 17 June 2009 - 04:20 PM

They should be OK, you might want to slow the acceleration (dw - delta omega) a bit though