Posted 18 March 2009 - 01:32 AM

I did a maxi-prep using the kit from QIAGEN. I then did a 1000 times dilution of the DNA and did a reading using quartz cuvettes in my ultraspectrophotometer. The reading showed a good yeild of DNA, about 3.2ug/ul.
I then did the old fasioned way of gauging the DNA concentration  by doing a serial dilution of the sample and loading the various dilutions in  a gel with 10 ul of a DNA ladder. By comparing the DNA concentration of a reference band in the ladder with the various dilutions, my DNA content is actually possibly a forth of what the spectrophotometer told me it was.

Anybody got ideas as to why it is like that. The purity of my sample as measured by the spectrophotometer was okay.



PLz Help

#2 HomeBrew

HomeBrew

    Veteran

  • Global Moderators
  • PipPipPipPipPipPipPipPipPipPip
  • 927 posts
14
Good

Posted 18 March 2009 - 02:48 AM

My quick reply would be to read the title of your question as the answer to your question, as a general rule.  There are many things that can throw your spec readings off...

#3 mdfenko

mdfenko

    an elder

  • Active Members
  • PipPipPipPipPipPipPipPipPipPip
  • 2,245 posts
78
Excellent

Posted 18 March 2009 - 10:36 AM

do you blank the spec with the same solvent in which you dilute your dna?

do you dilute your dna with the same solvent that it is in?
talent does what it can
genius does what it must
i do what i get paid to do

#4 perneseblue

perneseblue

    Unlimited ligation works!

  • Global Moderators
  • PipPipPipPipPipPipPipPipPipPip
  • 530 posts
8
Neutral

Posted 18 March 2009 - 10:57 AM

View PostHomeBrew, on Mar 18 2009, 02:48 AM, said:

My quick reply would be to read the title of your question as the answer to your question, as a general rule.  There are many things that can throw your spec readings off...


In addition, the spectrophotometer gives you the total DNA in your sample, including DNA fragments which individually are too low in concentration to be observed on a gel.
May your PCR products be long, your protocols short and your boss on holiday

#5 bob1

bob1

    Thelymitra pulchella

  • Global Moderators
  • PipPipPipPipPipPipPipPipPipPip
  • 4,347 posts
224
Excellent

Posted 19 March 2009 - 12:17 AM

As said before, EtBr fluorescence is not very sensitive, and will not pick up minor fragments that will be picked up by a spec.

The gel thing only works if you can load a serial dilution of the ladder, or even better spot (not run) serial dilutions of a known quantity of DNA onto a gel/sheet of plastic with dilutions of your sample side by side and then run a standard curve off the brightness of each spot.

Trying to estimate off one known point is completely useless.

#6 MaggieRoara

MaggieRoara

    Enthusiast

  • Active Members
  • PipPipPipPipPip
  • 67 posts
0
Neutral

Posted 19 March 2009 - 01:54 AM

View Postbob1, on Mar 19 2009, 12:17 AM, said:

As said before, EtBr fluorescence is not very sensitive, and will not pick up minor fragments that will be picked up by a spec.

The gel thing only works if you can load a serial dilution of the ladder, or even better spot (not run) serial dilutions of a known quantity of DNA onto a gel/sheet of plastic with dilutions of your sample side by side and then run a standard curve off the brightness of each spot.

Trying to estimate off one known point is completely useless.


How do I carry out your method? i gotta mix EtBr with the DNA then spot it onto a gel/ sheet of plastic?

#7 bob1

bob1

    Thelymitra pulchella

  • Global Moderators
  • PipPipPipPipPipPipPipPipPipPip
  • 4,347 posts
224
Excellent

Posted 19 March 2009 - 03:32 PM

View PostMaggieRoara, on Mar 19 2009, 01:54 AM, said:

View Postbob1, on Mar 19 2009, 12:17 AM, said:

As said before, EtBr fluorescence is not very sensitive, and will not pick up minor fragments that will be picked up by a spec.

The gel thing only works if you can load a serial dilution of the ladder, or even better spot (not run) serial dilutions of a known quantity of DNA onto a gel/sheet of plastic with dilutions of your sample side by side and then run a standard curve off the brightness of each spot.

Trying to estimate off one known point is completely useless.


How do I carry out your method? i gotta mix EtBr with the DNA then spot it onto a gel/ sheet of plastic?

Have a look in Sambrook, Fritsch and Maniatis, Molecular cloning, a laboratory manual.  The full method is in there.  Mix the DNA with an EtBr solution and then spot equal volumes onto a gel or sheet of plastic.  Fluoresce on a UV source and take a photo, then do some densitometry on the spots.  Be aware that saturation of the camera or photo will be reached at a relatively low level of DNA.