I have alot of spots on my western blot (looks like a starry sky) I know it cant be because of my anti mouse HRP conjugating cuz my beta actin control (which also uses the same anti mouse ) does not have the problem. Thinking it might be because my membrane might have been dirty prior to using it for transfer, i stripped the beta actin control membrane, inucubated it for a shorter secondary antibody incubation time (1hr; previously it was 1hr 30 minutes) but this time I have no bands at all, but the starry sky is still there. I have stripped again, now i am incubating with a lower Ab concetration of 1:10,000 instead of previous 1:1000. Havent developed yet.
Have any of you guys had the same problem?
Western blot has alot of spots
Started by MaggieRoara, Mar 08 2009 10:09 PM
5 replies to this topic
#1
Posted 08 March 2009 - 10:09 PM
#2
Posted 09 March 2009 - 12:41 AM
I remember that someone has such problem. He solved the problem by preparing new SDS solution.
#4
Posted 09 March 2009 - 04:04 AM
any possibility that they might be bubbles trapped between gel and the membrane during electroblotting?
#5
Posted 09 March 2009 - 04:09 AM
If I understand well, you perform WB anti-beta actin : all is fine,
if you strip this membrane and blot with your antibody of interest, you have a starry sky.
The same with a membrane freshly transferred and blotted with your antibody of interest.
if yes, don't lose time with preparing new SDS, it is not responsable for your problems.
what do you use as blocking reagent? milk?
do you blot against phospho-protein? biotinylated protein?
if you strip this membrane and blot with your antibody of interest, you have a starry sky.
The same with a membrane freshly transferred and blotted with your antibody of interest.
if yes, don't lose time with preparing new SDS, it is not responsable for your problems.
what do you use as blocking reagent? milk?
do you blot against phospho-protein? biotinylated protein?
Edited by little mouse, 09 March 2009 - 04:11 AM.
#6
Posted 13 March 2009 - 09:08 PM
I have similar problem but only with 2 antibodies for caspase-10 and GRP78. for many other antibodies (over 25) I do not have that problem. It seems to be related to the the antobody source.