Hey guys!
I'm new here. I wanted to determine concentration of my antibody. do you guys think i can use nanodrop and measure the absorbance at 280nm. if i use BCA assay, can i use BSA as my protein standard or should i use bovine gamma globulin as my standard. my budget is limited so i was thinking to use BSA. is it ok? 
From my experience, Nanodrop is not working reliable with protein 280nm and for BCA, Bradford, etc. you should use cuvettes anyway since the samples are not suitable for small volumes.
If you want to work in small volumes (you can do that with purified proteins like antibodies) you should use a NanoPhotometer from Implen, it is perfect for protein samples since their technology is not depending on surface tension of your samples. BCA, Lowry, etc. is also possible since the instrument works with cuvettes, too.
By the way, BSA is not a good standard for antibodies, since the molar extinction coefficient is far away from your antibody samples. You may want to use the molar extinction coefficient calculated from databases like
http://www.expasy.ch/ so you would be able to determine your 280 factor without any standard.
Hope I could help!