Fluorescence & Flow Cytometry Tutorial
http://www.invitrogen.com/site/us/en/home/.../Tutorials.html
Fluorescence & Flow Cytometry Tutorial
Started by Minnie Mouse, Jan 27 2009 01:41 PM
6 replies to this topic
#1
Posted 27 January 2009 - 01:41 PM
#2
Posted 27 January 2009 - 05:07 PM
Thank you, Minnie
Simple living, highnot thinking
#3
#4
Posted 07 March 2010 - 11:46 PM
if any body there can tell about how to calculate relative fluorescence units out of his to gram analysis in FACS. I want to know about mean, geometric mean and CV and their conversion into more understandable format of RFU and SD?
#5
#6
Posted 22 March 2010 - 05:29 AM
gradbiotech, on Mar 8 2010, 09:46 AM, said:
if any body there can tell about how to calculate relative fluorescence units out of his to gram analysis in FACS. I want to know about mean, geometric mean and CV and their conversion into more understandable format of RFU and SD?
I don't think converting the mean fluorescence as measured by the flow cytometer to RFU is possible, or rather, would make sense. As the name indicates, realtive fluorescence units are no really actual, universally defined units of measurement, but merely indicate the relative intensity of the measured fluorescence as compared to each other. The mean fluorescence measured by the flow cytometer, on the other hand, is the mean of the integrated areas of all light pulses measured by the detector, and is possibly instrument specific.
TL;DR I think that RFU and mean fluorescence are instrument- and experiment-specific values which are meaningless without controls to which they can be compared, and can therefore not be converted. Please correct me if I'm wrong.
#7
Posted 02 February 2011 - 05:34 AM
gradbiotech, on 07 March 2010 - 11:46 PM, said:
if any body there can tell about how to calculate relative fluorescence units out of his to gram analysis in FACS.
First, you need calibration beads with known number of molecules of the fluorochrome of interest binded.