Hey,
I am analyzing the methylation status of two regions: one ecompasses 500bp, the other is around 800bp. Could I just designed primers at flanking sites and amply these two regions? I found usually only 200-300bp were used to amplification for Bisulfite sequencing in most referecne. Is there any PCR length limitation for this method.
Thanks a lot ,
Jimmy
Is there a PCR length limitation after Bisulfite treatment?
Started by jimmyhj, Nov 29 2004 12:16 PM
3 replies to this topic
#1
Posted 29 November 2004 - 12:16 PM
#2
Posted 29 November 2004 - 07:13 PM
I have not tried PCR size bigger than 450 bp but have seen in one paper that over 1 kb was amplified from bisulfite modified DNA. I doubt about it.
#3
Posted 20 December 2004 - 07:54 AM
I follow the advice of most authors that 300 bp is about as large as one can go because bisulfite modified DNA is very fragile. You may want to do a nested PCR to increase product-- your first product might be 500 or 800 bp but you might not see a visible band. Also it's ok to just know the status of 300 bp or so if you are seequencing a CpG island-- it will give you a taste for what you are looking at and you can amplify the rest of the region separately if you think you have something significant.
#4
Posted 23 December 2004 - 08:52 AM
jimmyhj, on Nov 29 2004, 01:16 PM, said:
Hey,
I am analyzing the methylation status of two regions: one ecompasses 500bp, the other is around 800bp. Could I just designed primers at flanking sites and amply these two regions? I found usually only 200-300bp were used to amplification for Bisulfite sequencing in most referecne. Is there any PCR length limitation for this method.
Thanks a lot ,
Jimmy
I am analyzing the methylation status of two regions: one ecompasses 500bp, the other is around 800bp. Could I just designed primers at flanking sites and amply these two regions? I found usually only 200-300bp were used to amplification for Bisulfite sequencing in most referecne. Is there any PCR length limitation for this method.
Thanks a lot ,
Jimmy
Use high quality kits to extract DNA and the fragile problem following bisulfite modification shoudl be decreased and then it should be possible to amplified lagger fragments compared with published papers.
Good luck.