ELISA edge effect


  • Please log in to reply
6 replies to this topic

#1 elisa2013

elisa2013

    member

  • Active Members
  • Pip
  • 7 posts
0
Neutral

Posted 08 October 2013 - 08:15 AM

What is the most effective approach to minimize edge effect in a 96well elisa? All reagents & plates are brought to room temperature before using. Plates are covered during incubations, stacking no more than 4 high. Coating takes place at 37C O/N in a humidified container. All other incubations take place at RT (all 1 hour incs.), but not in a humidified container, just on the benchtop with lids on the plates. Room temperature is consistant. Substrate is diethanolamine, not incubated in the dark. Plates are medium bind Greiner. Edge effect happening is a larger than expected drop from the top row to the second row in a 2-fold serial dilution down the plate, ends up more like 2.5-3 fold, other dilutions below the 2nd well come down as expected compared to the well above it.

1- Can coating at a shorter incubation(5 hours) period at 37 help or keep O/N inc with 4C instead? ODs may not be high enough, this is a lower binding assay to begin with, many time only getting 1 or 2 usable ODs for a sample-which makes it unreliable to quantitate when the edge effect happens.

2- Can using a high bind plate help, so that incubations might be shorter?

3- Can using a rotator for the plates during incubations help?

 

thank you



#2 Sotirios

Sotirios

    member

  • Active Members
  • Pip
  • 23 posts
0
Neutral

Posted 08 October 2013 - 08:53 AM

Edge effect is a big problem. There will be no difference between medium binding and high binding. By using a rotator, there will be a litlle change.

If you see gradient, check humidity during the washes or the step of filling the chromogen. If there is classic edge effect, the problem is mainly the coating step. Probably, there is the deed of increasement of the coating reagent concentration.


Athanasiou Sotirios

#3 Missle

Missle

    Veteran

  • Active Members
  • PipPipPipPipPipPipPipPipPipPip
  • 132 posts
18
Good

Posted 09 October 2013 - 04:01 AM

Shaking the plates is a good idea.  It will add consistency in your results and help to normalize the wells.  How many plates are you running at one time?



#4 elisa2013

elisa2013

    member

  • Active Members
  • Pip
  • 7 posts
0
Neutral

Posted 10 October 2013 - 08:39 AM

Typically about 12 plates at a time. They are all timed in separate batches of 4 plates, so inc times are not an issue. Reagent is added almost immediately after washing each plate to minimize drying out.

I tried a comparison of stacking only 2 plates high & added the humidified chamber (plastic box w/ wet paper towels) for all of the incubations & it still happened, I think slightly worse if any change at all.

We don't have a rotator, but heard it helps with consistancy.

Maybe I'll try 4C coating, decreasing 37C coating time &/or increasing Ag conc. for the coat.



#5 HieuDang

HieuDang

    member

  • Members
  • Pip
  • 1 posts
0
Neutral

Posted 14 January 2014 - 02:16 PM

Hi everybody

My question may not relate to the topic much but i am confusing about it. I am going to do EILISA and the serum samples have been kept in -80 C. But today, i forgot and kept it in room temperature around 2 hours. These samples are important for me so I wonder that my mistake can affect to ELISA results or not. I hope receive your advise as soon as possible. Thank you very much?



#6 mdfenko

mdfenko

    an elder

  • Active Members
  • PipPipPipPipPipPipPipPipPipPip
  • 2,735 posts
125
Excellent

Posted 15 January 2014 - 05:31 AM

Hi everybody

My question may not relate to the topic much but i am confusing about it. I am going to do EILISA and the serum samples have been kept in -80 C. But today, i forgot and kept it in room temperature around 2 hours. These samples are important for me so I wonder that my mistake can affect to ELISA results or not. I hope receive your advise as soon as possible. Thank you very much?

it's possible that some proteases may degrade some of the proteins in the sample.


talent does what it can
genius does what it must
i do what i get paid to do

#7 tr31

tr31

    member

  • Active Members
  • Pip
  • 5 posts
0
Neutral

Posted 11 June 2014 - 12:30 PM

Hello everyone

So what kits would you people recommend for doing a sandwich ELISA protocol.

I am trying to look for IL-1 beta cytokine from cell supernatants.






Home - About - Terms of Service - Privacy - Contact Us

©1999-2013 Protocol Online, All rights reserved.