Leaving your polymerase at room temperature and exposing it to light will have an affect the polymerases stability. If you leave it in the light, it has the potential to photobleach your fluorescent pigment. Leaving your fluorescent dye on ice should not be a problem. I was referring to a control sample that you know shows good amplification in the past. Preferably a sample that uses the same primers are your experimental sample. Primer degradation could alter your Ct values. If your primer begins to degrade, small nucleotide sequences can bind, replicate and give you small fluctuations (probably not your cause, you have minimal fluctuations). Your samples are measured by their fluorescence. If your plate is warped, it you have a bubble formations or a smudge on your optical film, this will hinder the machine to take an accurate reading. This will cause small deviations in your fluorescence reading each round of replication. Whenever I finish loading my plate, I spin it down and give the optical film a quick wipe. Who knows if something inadvertently got on your optical film.
I was curious to see if you ran this sample in doublets because I wanted to know if this was a average of the two samples. If this is not an average, did you other sample show this same fluctuation?
After you finish your qPCR, run your samples on a gel. Are you still getting a sharp band from when you initially tested your primers and polymerase at the current cycling conditions.
In all honesty. Your curve is good. The relative fluctuations are probably just the result of the Sybr green being uncooperative.