Stain in electrophoresis is going from + to -


  • Please log in to reply
3 replies to this topic

#1 Gwynbleidd

Gwynbleidd

    member

  • Members
  • Pip
  • 3 posts
0
Neutral

Posted 22 May 2013 - 09:48 AM

Hey, I have small problem during DNA electrophoresis.
I am using Serva Dna Stain G. So I have to run gel quite long because I have 2 framgents (one is much bigger than the other one) but when I am running a gel (1%) after 45 min at 75V the stain is going up (like almost to half of it) so then I can't see any bands below. So I the stain is negativly charged.
As I remember with EtBr I have not such a problems? Or my memory is messing?

So the only think that I can try is post-staning? Or there is something else?
of course I can also do a extremely long gel :)

#2 bob1

bob1

    Thelymitra pulchella

  • Global Moderators
  • PipPipPipPipPipPipPipPipPipPip
  • 5,786 posts
406
Excellent

Posted 22 May 2013 - 12:44 PM

You should be able to add some stain to the running buffer.

#3 Gwynbleidd

Gwynbleidd

    member

  • Members
  • Pip
  • 3 posts
0
Neutral

Posted 22 May 2013 - 02:31 PM

with what concentration? The same as in the gel? but it is little bit wasting of stain.
Just for curious with EtBr is the same problem?
and how good is post staining?

#4 phage434

phage434

    Veteran

  • Global Moderators
  • PipPipPipPipPipPipPipPipPipPip
  • 2,484 posts
251
Excellent

Posted 22 May 2013 - 03:44 PM

EtBr has the same problem. Either stain after running the gel, or add stain to the buffer at the positive electrode.