Cause of random samples failing PCR?

PCR Genomic DNA Molecular Biology

  • Please log in to reply
2 replies to this topic

#1 Epigeneticist

Epigeneticist

    Enthusiast

  • Active Members
  • PipPipPipPipPip
  • 97 posts
5
Neutral

Posted 03 May 2013 - 09:50 AM

When running various PCR reactions I get random samples that fail while other samples amplify fine. I re-run these samples and they work fine. On failure, they produce a band of low molecular weight (100-200 bp) which sometimes looks like a smear with complete absence of my desired product size or my desired band is faint. For example, I will run 10 samples and 9 or 10 will work but one will fail. I am not using a hot start polymerase. However, my reaction mixture and reaction tubes are kept on ice at all times. I do pipette my mastermix up and down a 5-10 times before aliquoting which would be the only time other than aliqouting that the mastermix has a chance to briefly warm up. After adding gDNA to the reaction tube I do not mix, I just pipette the small volume of gDNA up and down a few times. I also have my thermal cycler pre-warmed.

Any explanations for the random failures?

Mastermix mixing problem? Should I vortex my mastermix lightly or pipette up and down before aliquoting?

Not mixing sample into reaction enough? Should I give them a slight vortex?

How sensitive is a non hot start polymerase to brief temperature increase (i..e few seconds during mixing or transferring to thermal cycler) ?

Thanks

I attached a picture. Top gel is what I am referring to and bottom gel too. Bottom gel also has an issue with the non-specific high MW bands. DNA ladder is 100bp with the brighest band = 500bp. The desired bands are supposed to vary in size because it is a genotype assay.

Attached Thumbnails

  • Untitled.jpg


#2 Ameya P

Ameya P

    Rervm Cognoscere Cavsas

  • Moderators
  • PipPipPipPipPipPipPipPipPipPip
  • 333 posts
26
Excellent

Posted 05 May 2013 - 11:51 PM

Hi Epigeneticist,

There could be a couple of reasons why your PCR fails for a few samples.
  • SInce you are adding template DNA from various sources, their DNA concentration could be variable.
  • If you are using lysis solutions for your extraction, salt remnants can also impact how well your PCR works out.
  • Also, you said that you DO NOT mix your gDNA into your master mix. Well, I always mix it.
  • How gDNA do you add? Porbably would be a good idea to get you pipette calibrated if you pipette volumes that are less than 2 ul

Hope this helps.

Ameya

NEW!!!!  Wonder why this was not part of Chemistry class, ever!  on CoffeeTableScience!!!!

Image copyright: Adrian Koh SF.
Replication of this art is strictly prohibited without express permission of the artist


#3 Epigeneticist

Epigeneticist

    Enthusiast

  • Active Members
  • PipPipPipPipPip
  • 97 posts
5
Neutral

Posted 13 May 2013 - 05:33 PM

Hi Epigeneticist,

There could be a couple of reasons why your PCR fails for a few samples.

  • SInce you are adding template DNA from various sources, their DNA concentration could be variable.
  • If you are using lysis solutions for your extraction, salt remnants can also impact how well your PCR works out.
  • Also, you said that you DO NOT mix your gDNA into your master mix. Well, I always mix it.
  • How gDNA do you add? Porbably would be a good idea to get you pipette calibrated if you pipette volumes that are less than 2 ul

Hope this helps.

Ameya


Thanks for the response Ameya and your suggestions. My problem is due to the polymerase which is not a hot start enzyme. I was doing a poor job of keep my samples on ice during setup. I have been very careful to keep everything on ice and I haven't had any issues.

I ruled out the problems you suggested based on the following:
DNA concentration - I check my samples in duplicate and I even verified my diluted samples too.
Salt in DNA extractions - extractions are usually very good, and DNA eluates are concentrated so I have to dilute them which would dilute out the salt if I had any.
Pipettes calibration - calibrated by a certified company

Thanks again for your response! Do you have any explanation for the cause of non-specific high molecular bands (between 600-1000bp) or how to get rid of them? I have read too much template could cause this but I do not think 100ng of genomic DNA is excessive. My product is only 400bp so I decreased my extension time from 1min to 30 seconds and it got rid of the bands. Just wondering if there is another way to approach it.