I used ELISA kit that provide microtiter plate, 96 wells precoated with Human N-MID-OT polyclonal antibodies.
the kit is very expensive and is that possible if I wash the plate and re used the plate?
Does the precoated antibody already affected?
Submit your paper to J Biol Methods today!

Re use the pre coated microtiter plate 96 wells
Started by hidayahcellculture, Apr 22 2013 01:16 AM
2 replies to this topic
#1
Posted 22 April 2013 - 01:16 AM
#2
Posted 22 April 2013 - 01:31 AM
you can find a similar discussion here- http://www.protocol-...osts/39925.html
I always had an alternate hypothesis....
#3
Posted 22 April 2013 - 05:16 PM
It certainly would not be a recommended approach to re-use an ELISA plate from a kit. If the reagents are stripped off using acid (1 M acetic acid for example) there may be some binding activity left, but the assay performance is likely to be changed so that the dynamic range is likely to be shifted and precision is likely to be widened. If the kit is obtainded from a reputable manufacturer the cost of the kit is a reflection of the manufacturer's research and effort put into developing the kit so that you can provide quality research data. The dollars that you might save by trying to reuse these plates will work against the quality of your research.