qPCR: band in gel, but super high Cq?
#1
Posted 12 November 2012 - 07:19 AM
I'm running RT-qPCR for relative expression on tissues.
I ran some testsamples on with conventional PCR (cDNA 1:10) and looked at them on a gel. In comparison to my reference gene the bands look alright and I expected them to give me nice Cqs. Also used 4x the cDNA in the testrun, but the bands on the gel actually didn't look different.
Ran the lot of my samples in the qPCR then, but only get Cqs of 37-39.
I am using the Roche UPL system, so the amplicon is detected by a probe that is specific.
Any ideas what could be the problem?
Thanks!
#2
Posted 12 November 2012 - 11:21 AM
It could also be that you have some sort of inhibitors in the cDNA (carried over from extraction) that are inhibiting the PCR itself, these might be diluted out sufficiently in a 1:10 dilution such that the PCR works better.
#3
Posted 12 November 2012 - 01:22 PM
Did you looked at the gel of the product from the real-time reaction?
And actually in UPL system probes are not specific, only primers are.
I never trust anything that can't be doubted.
#4
Posted 13 November 2012 - 02:01 AM
bob1, on 12 November 2012 - 11:21 AM, said:
I see. Good to know. Thanks
Quote
Nanodrop values are superb. 260/280 is 2.15 in average and 260/230 is 2.27 in average for my samples
I'm using RNeasy Lipid extraction kit, which uses Trizol in combination with spin columns.
I elute RNA in 60ul H20. Plenty of good quality RNA there.
Quote
Did you looked at the gel of the product from the real-time reaction?
And actually in UPL system probes are not specific, only primers are
I use the same chemicals for conventional (minus the probe) and real-time PCR. Only the thermocycler and tubes/plates are different. I also put some of my qPCR samples onto the gel, which shows a band just like the conventional PCR.
Edited by Podge, 13 November 2012 - 02:43 AM.
#5
Posted 13 November 2012 - 09:19 AM
Question is, whether UPL system works in your conditions at all (maybe noncompatible mastermix or something) or whether this particular assay is bad. Because they sometimes can be.
Are you using Roche mastermix, cycler?
Did you had ever a good working UPL assay or is this your first? If you did, then I would just design new primers/assay with a different probe if you have probe set.
If it's your first time (or you have several assays that have same problem), maybe I would try to optimise, but the only thing that you really can if your reaction runs fine is probably concentration..
Also check you are using the right channel (FAM).
I never trust anything that can't be doubted.
#6
Posted 13 November 2012 - 11:54 AM
Podge, on 13 November 2012 - 02:01 AM, said:
Quote
Nanodrop values are superb. 260/280 is 2.15 in average and 260/230 is 2.27 in average for my samples
I'm using RNeasy Lipid extraction kit, which uses Trizol in combination with spin columns.
I elute RNA in 60ul H20. Plenty of good quality RNA there.
#7
Posted 14 November 2012 - 03:58 AM
Trof, on 13 November 2012 - 09:19 AM, said:
Question is, whether UPL system works in your conditions at all (maybe noncompatible mastermix or something) or whether this particular assay is bad. Because they sometimes can be.
Are you using Roche mastermix, cycler?
Did you had ever a good working UPL assay or is this your first? If you did, then I would just design new primers/assay with a different probe if you have probe set.
If it's your first time (or you have several assays that have same problem), maybe I would try to optimise, but the only thing that you really can if your reaction runs fine is probably concentration..
Also check you are using the right channel (FAM).
Hey.
Yes. Band on gel, but no signal/very very late signal in qPCR.
I always design the primers using the design centre and use one of the top5 that matches all prerequesites. Saying that I have ran many a UPL assay. So it is not the problem of mastermix or the real-time machine. I do you corresponding reference gene values for my samples, so it has to be in the primers or probe.
By "If it's your first time (or you have several assays that have same problem), maybe I would try to optimise,..." you mean optimising primer concentration etc.
I have now designed new primers and ordered them.
Quote
Good to know the value should be between 1.9-2.0. I somehow always thought a perfect 260:280 ratio is 2.3!
Here a screenshot of my 16 samples. Apart from one with a crappy value (light blue), all of them were good.
I will run some RNA samples on a gel later. Do you know how much one adds to a gel normally?
#8
Posted 14 November 2012 - 01:19 PM
Degraded RNA and DNA will still give you a reading on the nano-drop, even nucleotides will!
#9
Posted 14 November 2012 - 02:01 PM
Podge: Optimising.. well, primers maybe, but I'm not sure what effect it would have on the binding of the probe, by running the product on a gel you proved your PCR is running fine, so maybe optimising probe concentration or something (or maybe even the primers/probes ratio, since there may be some interaction). They claim I think 96% success rate with one of first three primers designed by Design Center, so sometimes you fall into that 4 %.
In case a UPL design fails I order another, if that's not possible for some reason I just use the primers, which are good and run the assay on SYBR Green. UPL is not designed to be optimised much.
I never trust anything that can't be doubted.
#10
Posted 14 November 2012 - 11:08 PM
Trof, on 14 November 2012 - 02:01 PM, said:
#11
Posted 22 November 2012 - 06:08 AM
sorry for the late answer, but was on some holidays for an extended weekend.
Here are my results with this particular PCR.
UPL:
cDNA 1:2.5: Cq ~35
cDNA 1:10: Cq ~37-39
Gel: right band at 132bp
SybrGreen:
cDNA 1:2.5: Cq ~32
cDNA 1:10: Cq ~28
Gel: primer dimers appear
Product not the right size (200bp instead of 132bp)
- Different size for 1:10 and 1:2.5!!
Slightly different Tm for 1:10 and 1:2.5 samples
All in all it seems like the whole Primer plus Probe combination is messed up. Dimers. different band sizes?! Got new primers today and will test them.
At least I know the gene is there in some way. Tried optimising this PCR in a different tissue where we weren't sure about expression levels. At least I know now it was the assay itself.