positive NTC in real time reaction
#1
Posted 03 September 2012 - 12:41 AM
In last RT reactions i get a positive NTC (~cycle 30). The strange thing is that I ussually load two NTC samples in each reaction, and one NTC gives no signal, but the second one, which is following my samples on plata, gives a signal. My lowest samples ended on cycle 26-27.
Is this mean I have a NTC contamination? Please help me!!!
Thank you
Natali
#2
Posted 03 September 2012 - 02:29 AM
-- Bernard M. Baruch
#3
Posted 03 September 2012 - 03:13 AM
If the contamination was in the reaction premix, both NTCs would be possitive. This however means that you have improper handling of a pipette and you may have cross contaminated other sampes as well.
Make a very strict rules about the movement of the pippete above adjacent wells. Try to move between wells or completely outside when possible. Don't release the knob unless you put the tip directly above the bin. Don't put new tip boxes in the trajectory that leads to bin. Don't splash or pipette up and down furiously.
It can still be normal contamination though, from the air, or the pipette, so generaly stick to GLP.
Run the reaction again, this time very carefully. If that still happen, clean everything and run only NTCs on a plate to check your reagents are OK. Putting NTC in wells far away from your samples would help NTC cross contamination from samples, but that would only hide the primary problem with the improper pipetting so I don't recomend that.
I never trust anything that can't be doubted.
#4
Posted 03 September 2012 - 08:41 PM
-- Bernard M. Baruch
#5
Posted 04 September 2012 - 12:10 AM
prabhubct, on 03 September 2012 - 08:41 PM, said:
I never trust anything that can't be doubted.
#6
Posted 05 September 2012 - 05:11 AM
I don't think it's an improper handling of a pipette, because I have this NTC read only resently (in last experimental set), before I had no NTC read at all. And, I do put my NTC in wells far away from the samples. Moreover, I close the tubes with NTC after adding the water to the mix even before I begin to add my samples. Maybe I have the contamination among the samples, but it not suppose to be in NTC.
#7
Posted 05 September 2012 - 10:04 AM
I never trust anything that can't be doubted.
#8
Posted 06 September 2012 - 02:55 AM
#9
Posted 06 September 2012 - 05:15 AM
Edited by prabhubct, 06 September 2012 - 05:16 AM.
-- Bernard M. Baruch
#10
Posted 06 September 2012 - 11:30 AM
#11
Posted 06 September 2012 - 12:57 PM
nicelady8, on 06 September 2012 - 02:55 AM, said:
If you checked also the length of the product on gel it is indeed contamination.
For that, it stays true what I wrote earlier. It's unlikely you've got contaminated reagents, because all NTCs would be positive. So.. cross contamination from pipetting, or contamination of well.
I never trust anything that can't be doubted.
#12
Posted 06 September 2012 - 11:07 PM
But, it still shows some amplification sometimes, its because of unused primer getting jumbled which migth somehow retain the SYBR green showing some kind of amplication, which is not real.
So, if you are sure about your techniques and every steps you have gone through, there is nothing to worry about the amplificaiton seen in the later cycles. Just like you mentioned above 28-cycles. Which is fairly acceptable.
Good luck with your research!
Have a great time ahead.
#13
Posted 07 September 2012 - 05:38 AM
But important is it only happens in the absence of template, and that it has a different melitng profile. I would certainly worry about amplification seen in later cycles if it had the same melting profile as a product, because 28 is not unusual value for some samples. It would not be acceptable, so it's important to add melting step after each amplification with SYBR, otherwise you don't know what exactly are you looking at in the amplification graph.
I never trust anything that can't be doubted.