Grainy image and no signal on Western using ChemiDoc


  • Please log in to reply
6 replies to this topic

#1 LabGirl

LabGirl

    member

  • Active Members
  • Pip
  • 17 posts
0
Neutral

Posted 07 June 2012 - 09:00 AM

We are having issues with our ChemiDoc for imaging Westerns and wondering if anyone has suggestions? The Western blot conditions have been optimized and we were getting consistent results previously. However, now we are getting grainy/spotted images using our ChemiDoc. It is happening to multiple users with multiple blots. It's only happening when we image Westerns; our agarose gels are not grainy, so we don't think its the camera or dust. We are using Immun-Star Western C Chemi Kit from biorad at 1000 ul total on each membrane. When the ChemiDoc is imaging, the pictures are black and no bands are seen, however when we go to review the image, we just see a grainy picture. An example image is attached.

Any suggestions would be helpful, thanks!!

Attached Thumbnails

  • Example blot.jpg


#2 mdfenko

mdfenko

    an elder

  • Active Members
  • PipPipPipPipPipPipPipPipPipPip
  • 2,735 posts
125
Excellent

Posted 07 June 2012 - 11:54 AM

i can think of a couple of possibilities:

the camera resolution is set wrong

the vessel in which you are processing the blot needs to be thoroughly cleaned

Edited by mdfenko, 07 June 2012 - 11:55 AM.

talent does what it can
genius does what it must
i do what i get paid to do

#3 LabGirl

LabGirl

    member

  • Active Members
  • Pip
  • 17 posts
0
Neutral

Posted 12 June 2012 - 01:18 PM

Hi mdfenko,

Thanks for the input. I am sure the camera resolution is fine-sometimes the Western blots do work (and work well), and the camera itself should be fine: we image agarose gels and other items on this ChemiDoc with no problems. We have been consistently cleaning very thoroughly the entire time, and at every step of the blot, but are still having the same problems. Any other options you can think of? I think it could simply be an antibody problem, but sometimes this antibody will work, sometimes not.

#4 mdfenko

mdfenko

    an elder

  • Active Members
  • PipPipPipPipPipPipPipPipPipPip
  • 2,735 posts
125
Excellent

Posted 15 June 2012 - 08:34 AM

what do you use to block?

maybe your block is sometimes not fully solubilized or has precipitated.

Edited by mdfenko, 15 June 2012 - 08:34 AM.

talent does what it can
genius does what it must
i do what i get paid to do

#5 LabGirl

LabGirl

    member

  • Active Members
  • Pip
  • 17 posts
0
Neutral

Posted 15 June 2012 - 12:23 PM

Hi mdfenko,

I thought the same thing, however we now know what was going on. I was posting this for my grad student, and apparently she had reduced the amount of the protein sample she was loading by more than 50% and failed to mention it, so I think that was the issue. I believe the grainy appearance was the imager trying to pick up a signal where there wasn't any. I suggested increasing the amount of protein she was adding and also using a positive control.

Thanks for the reply!

#6 science noob

science noob

    Veteran

  • Active Members
  • PipPipPipPipPipPipPipPipPipPip
  • 275 posts
20
Excellent

Posted 15 June 2012 - 01:09 PM

What is the exposure time for this image?

#7 Lynn Pais

Lynn Pais

    member

  • Members
  • Pip
  • 1 posts
0
Neutral

Posted 28 August 2013 - 07:32 AM

Hi, I'm having the exact same problem. I think its got something to do with the exposure time. Could really use some help here!