My experience with sequencing basically involves everything necessary to get DNA prepped for sequencing and then sending it away to get sequenced. I also have only worked exclusively with Drosophila. Now, I work for a start-up with limited funds, so I am trying to get this working without having to buy too many more reagents/machines.
Background: We are trying to sequence SNPs from human DNA
Amplicons range from 99bp to 500bp. We have used an appropriate Qiagen kit for extraction.
1. I do not have access to a spectrophotometer to determine DNA concentration, I have seen Picogreen recommended, however this still requires a spec. Can I just use a DNA ladder to estimate concentration or will this simply not be accurate enough? Otherwise we'd have to buy a spec + accessories + picogreen + reagents, etc...
2. After performing the PCR off my DNA template, can I use an ethanol precipitation to clean it up before running the Beckman Coulter DTCS kit (BC suggests the qiagen PCR clean-up kit)?
3. I know the CEQ 8000 can do SNP analysis, anyone have experience with this?
Any help would be greatly appreciated, as I do not have any co-workers or friends that have experience with this machine either! It's brand new to all of us.
New to sequencing, questions about Beckman Coulter CEQ 8000
Started by contactsports, Jun 06 2012 10:19 AM
sequencing beckman coulter ceq 8000
3 replies to this topic
#1
Posted 06 June 2012 - 10:19 AM
#2
Posted 06 June 2012 - 11:56 AM
we have a ceq 8000.
1. an accurate determination is desired. we've seen that it is best to use ~100 fmoles of template for best results, more may clog the capillaries, less may not yield enough product to get accurate results (despite what b-c says about optimum amount of template).
since you will be using pcr products as template (?), you can safely go over the limit by a small percentage, so you may be able to get away with comparing to the ladder.
2. the reason to use a pcr cleanup, rather than ethanol precipitation, is to eliminate the original template dna as well as the reactants. if you don't want to use the qiagen (or similar) spin column pcr cleanup then you'll have to clean the pcr product(s) on a gel. then you can ethanol precipitate it after extracting from the gel.
3. snp analysis on the ceq is performed by fragment analysis. the kit is different (you don't use dtcs). we've never performed snp analysis with this machine (we do perform fragment analysis).
1. an accurate determination is desired. we've seen that it is best to use ~100 fmoles of template for best results, more may clog the capillaries, less may not yield enough product to get accurate results (despite what b-c says about optimum amount of template).
since you will be using pcr products as template (?), you can safely go over the limit by a small percentage, so you may be able to get away with comparing to the ladder.
2. the reason to use a pcr cleanup, rather than ethanol precipitation, is to eliminate the original template dna as well as the reactants. if you don't want to use the qiagen (or similar) spin column pcr cleanup then you'll have to clean the pcr product(s) on a gel. then you can ethanol precipitate it after extracting from the gel.
3. snp analysis on the ceq is performed by fragment analysis. the kit is different (you don't use dtcs). we've never performed snp analysis with this machine (we do perform fragment analysis).
talent does what it can
genius does what it must
i do what i get paid to do
genius does what it must
i do what i get paid to do
#3
Posted 06 June 2012 - 01:16 PM
mdfenko, on 06 June 2012 - 11:56 AM, said:
we have a ceq 8000.
1. an accurate determination is desired. we've seen that it is best to use ~100 fmoles of template for best results, more may clog the capillaries, less may not yield enough product to get accurate results (despite what b-c says about optimum amount of template).
since you will be using pcr products as template (?), you can safely go over the limit by a small percentage, so you may be able to get away with comparing to the ladder.
2. the reason to use a pcr cleanup, rather than ethanol precipitation, is to eliminate the original template dna as well as the reactants. if you don't want to use the qiagen (or similar) spin column pcr cleanup then you'll have to clean the pcr product(s) on a gel. then you can ethanol precipitate it after extracting from the gel.
3. snp analysis on the ceq is performed by fragment analysis. the kit is different (you don't use dtcs). we've never performed snp analysis with this machine (we do perform fragment analysis).
1. an accurate determination is desired. we've seen that it is best to use ~100 fmoles of template for best results, more may clog the capillaries, less may not yield enough product to get accurate results (despite what b-c says about optimum amount of template).
since you will be using pcr products as template (?), you can safely go over the limit by a small percentage, so you may be able to get away with comparing to the ladder.
2. the reason to use a pcr cleanup, rather than ethanol precipitation, is to eliminate the original template dna as well as the reactants. if you don't want to use the qiagen (or similar) spin column pcr cleanup then you'll have to clean the pcr product(s) on a gel. then you can ethanol precipitate it after extracting from the gel.
3. snp analysis on the ceq is performed by fragment analysis. the kit is different (you don't use dtcs). we've never performed snp analysis with this machine (we do perform fragment analysis).
Thank you for responding so quickly! I figured the ethanol precipitation wouldn't be enough. Have you ever tried the Diffinity Rapidtip pcr purification method (http://www.diffinity...ts_rapidtip.asp)?
I ran a trial run throught the machine and got the puc18(?) control to work (trust me, that surprised me), so I am hopeful that I can get some results with the help you have provided me. I'll probably be back with more questions once I have the proper template DNA for the machine.
#4
Posted 07 June 2012 - 11:39 AM
contactsports, on 06 June 2012 - 01:16 PM, said:
Have you ever tried the Diffinity Rapidtip pcr purification method (http://www.diffinity...ts_rapidtip.asp)?
Quote
I ran a trial run through the machine and got the puc18(?) control to work (trust me, that surprised me), so I am hopeful that I can get some results with the help you have provided me. I'll probably be back with more questions once I have the proper template DNA for the machine.
Edited by mdfenko, 07 June 2012 - 11:39 AM.
talent does what it can
genius does what it must
i do what i get paid to do
genius does what it must
i do what i get paid to do
Also tagged with one or more of these keywords: sequencing, beckman coulter, ceq 8000
![]() |
Protocols and Techniques Forums →
Molecular Cloning →
problem with sequencing and mappingStarted by Guest_ahmadfathi_* , 27 Apr 2013 |
|
|
|
Protocols and Techniques Forums →
Molecular Biology →
How do I know the orientation of my insert in a plasmid after sequencing?Started by Guest_bioguy_* , 18 Apr 2013 |
|
|
||
Protocols and Techniques Forums →
Molecular Biology →
Has anyone reused any BigDye Cleanup plates?Started by Guest_Lefty_* , 24 Jan 2013 |
|
|
||
Protocols and Techniques Forums →
Molecular Biology →
Sanger Sequencing RNAse Interference?Started by Guest_Dino_* , 14 Jan 2013 |
|
|
||
Protocols and Techniques Forums →
DNA Methylation and Epigenetics →
Biotinylated primersStarted by Guest_Swongel_* , 10 Dec 2012 |
|
|