I am using quantitive Real Time PCR for detection of etiologic agents from patients samples. Our current target is 33 pathogens. One company has provided us plasmids to be used as positive control. For each target we need to run standard curve per quater month and use the slope (m) and Y-intercept © for quantification of our samples.
For each target we are getting different slope (m). My question is "what is the ideal range for slope of qPCR standard curve?"
2 replies to this topic
#1
Posted 31 May 2012 - 07:42 AM
#2
Posted 31 May 2012 - 02:34 PM
Surely it is 1.
#3
Posted 31 May 2012 - 06:31 PM
I just find it.. The ideal slope should be -3.3. But the range is -3.1 to -3.6. But my problem is as I am dealing with 33 targets for some the slope is above -3.6. As the slope indicate PCR efficiency, I am worried about efficiency of my qPCR.
Could it be due to any problem with positive plasmids?
Could it be due to any problem with positive plasmids?
Also tagged with one or more of these keywords: qPCR, Standard Curve, TaqMan
![]() |
Protocols and Techniques Forums →
Tissue and Cell Culture →
Housekeeper for HEK cells?Started by Guest_Stephp_* , 17 May 2013 |
|
|
|
![]() |
Protocols and Techniques Forums →
PCR, RT-PCR and Real-Time PCR →
FAM and VIC S-curvesStarted by Guest_squallweathered_* , 23 Apr 2013 |
|
|
|
Protocols and Techniques Forums →
PCR, RT-PCR and Real-Time PCR →
using cell line to generate standard curveStarted by Guest_detriar_* , 14 Apr 2013 |
|
|
||
Protocols and Techniques Forums →
siRNA, microRNA and RNAi →
miRNA quantificationStarted by Guest_dude3005_* , 18 Mar 2013 |
|
|
||
Protocols and Techniques Forums →
PCR, RT-PCR and Real-Time PCR →
Confused about normalizationStarted by Guest_SillySaru_* , 14 Mar 2013 |
|
|