I was preparing 1.5M Tris-Hcl pH 8.8, and the solution is cloudy, which I don't understand why. I tried to make it with distilled water & MilliQ H2O, the same. I tried to increase the temperature to 35deg C and dissolve, but still the same. pH was maintained at 8.8 throughout. I also put the solution under stirrer for a whole night and still the same. I didn't add any NaOH at all.
When I used this Tris-Hcl to run my buffer, the electrophoresis was fine. And even the Western Blot turned out pretty good.
But I don't understand the reason for the buffer to be still cloudy. Can anyone explain? Thanks.
Submit your paper to J Biol Methods today!

Tris-Hcl becomes cloudy.
Started by VJbiofication, Feb 07 2012 03:06 AM
SDS PAGE electrophoresis Buffers Tris
3 replies to this topic
#1
Posted 07 February 2012 - 03:06 AM
#2
Posted 07 February 2012 - 12:56 PM
There could be a number of reasons - 1) it could be that your bottle(s) are dirty and something (perhaps detergents) are coming off into solution. 2) it could be that there is a contaminant in the Tris or (less likely) the HCl. 3) it could be that your MilliQ filters need to be changed as something is coming through. I have seen this before when someone cleaned the distiller for the water that supplied the MilliQ by rubbing off the calcium build up and not flushing the line before re-connecting it to the supply.
#3
Posted 08 February 2012 - 02:58 AM
may be therevis contamination
#4
Posted 08 February 2012 - 02:58 AM
may be there is contamination