Weird qPCR curves
#1
Posted 17 January 2012 - 07:11 PM
I have been doing qPCR for a while and everything was fine until last week. I observed this weird curve. Basically the SYBR signals dropped around cycle 30 for all my wells. I have never seen this before. I checked the machine and it was fine. I repeated the same experiment, still saw the same weird curve. The dissociation curve looks normal for both runs.
I can't figure out what went wrong. Has anyone seen similar curves before? Desperate for an answer. Any guess is welcome as well.
Thanks a lot guys!!!!
#2
Posted 18 January 2012 - 04:51 PM
#3
Posted 18 January 2012 - 06:06 PM
#4
Posted 18 January 2012 - 06:14 PM
Thanks a lot for your reply. My program was the same for 40 cycles. Okay, the overall program is 95oC 10min for 1 cycle, then 95oC 30s & 48oC 30s (40 cycles), and 1 cycle of 95oC 1min, 55oC 30s and 95oC 30s.
The liquid in the well was not evaporated because I checked for that.
It is very weird because it does not occur every single time. Sometimes it happens and sometimes it does not. It is driving me crazy as sometimes I have to pray before I do my experiments.....
phage434, on 18 January 2012 - 06:06 PM, said:
#5
Posted 18 January 2012 - 06:45 PM
#6
Posted 18 January 2012 - 06:50 PM
phage434, on 18 January 2012 - 06:45 PM, said:
Yes, I am using a heated lid. We use stratagene mx3000p for qPCR, which should have heated lid. Also I did not see evaporation around cycle 30 of my qPCR (I opened the lid and checked). My reaction volume is 20ul.
#7
Posted 18 January 2012 - 06:57 PM
#8
Posted 18 January 2012 - 07:11 PM
Edited by peanut142, 18 January 2012 - 07:12 PM.
#9
Posted 19 January 2012 - 11:04 AM
I'm trying to find the brochure where I read this, it was Roche I think.]
Hey, Trof, thanks for your suggestion. The weird curve does not affect wells with high copy number of candidate DNA, but messes up those which have low copy number, as the curve does not come up at the right cycle. Could you please share me the link if you find the brochure? I cannot find it myself.
Many thanks again!
Edited by peanut142, 19 January 2012 - 11:05 AM.
#10
Posted 19 January 2012 - 11:23 AM
This looks almost like kind of a machine error. But that would happen randomly in any experiment. If it's only in a particular assay, it's weird. Did you try to check it on gel later? SYBR shows decrease in DNA, but you could compare on gel if it's true. If not, then it's problem in SYBR (degrading?) or fluorescence measurement. I would probably try to send this to Stratagene support too, what they think.
phage434: 95/55/95 would be melting analysis I guess, there is missing the ramp from 55 to 95.
I never trust anything that can't be doubted.
#11
Posted 19 January 2012 - 11:33 AM
Trof, on 19 January 2012 - 11:23 AM, said:
This looks almost like kind of a machine error. But that would happen randomly in any experiment. If it's only in a particular assay, it's weird. Did you try to check it on gel later? SYBR shows decrease in DNA, but you could compare on gel if it's true. If not, then it's problem in SYBR (degrading?) or fluorescence measurement. I would probably try to send this to Stratagene support too, what they think.
phage434: 95/55/95 would be melting analysis I guess, there is missing the ramp from 55 to 95.
Thanks again. I will email our stratgene rep here.
#12
Posted 21 January 2012 - 04:33 PM
..."best of our knowledge, as far as we know this had never been reported before, though I can't possible read all the published journals on earth, but by perform thorough search in google, the keywords did not match any documents"...
"what doesn't kill you, makes you stronger"---Goddess Casandra reminds me to be strong
"It's all just DNA. Do it."---phage434