I have two questions, first:
I have recently been developing a protocol for western blotting all tyrosine phosphorylated proteins from B cell whole cell lysates. I have optimized the antibody concentrations and incubation times and now I am having a problem of inconstancy. Basically the band intensity seems to depend on where it is on the blot.
Using cleared whole cell lysates, frozen stocks.
Determined total protein concentration with BCA assay, serial dilution of BSA standard in identical lysis buffer
Loaded equivalent total protein on gel in Laemelli, determined by results of BCA assay
Loaded identical samples in duplicate, on opposite sides of gel
Transferred in OWL system, 260 mA for 1 hour, confirmed minimal blowthrough using second membrane
Blocked overnight in 4% BSA at 4 deg C, stained with primary for 3 hours at RT, and secondary 1 hour at RT, using PICO ECL.
All the bands on one side clearly had a higher intensity than the identically loaded bands on the other side.
Some things I am considering:
1) Transfer apparatus is transferring unevenly
2) antibody incubation needs more buffer to evenly cover blot (I don't think this is it because the blot is essentially floating and freely moving during agitation
3) the ECL substrate is not agitated while incubated. Could this be the issue? Are the blots commonly agitated while incubating with ECL reagent?
I am hoping to either reprobe the blots with different antibodies, or use a Ponceau stain to see if the protein amounts are equal. If anyone has any other ideas on how to determine the cause and or directly fix the problem, I would love to hear. Thanks.
secondly:
I want to look at the B cells at precise timepoints after stimulation. This is impossible using my current method of spinning down stimulated cells and resuspending in 1x RIPA buffer with phospho and protease inhibitors. Do people use concentrated lysis buffers to directly lyse suspension cells in media/other buffer instead of spinning into lysis buffer? Any input is greatly appreciated!