Gel gets squiggly bands for low molecular weight proteins


  • Please log in to reply
3 replies to this topic

#1 JesseC

JesseC

    member

  • Active Members
  • Pip
  • 10 posts
0
Neutral

Posted 10 March 2011 - 05:57 AM

Hi everyone, I'm new here and am just starting in a lab and have run into a problem that is driving me up the wall.  We are trying to look at proteins around 14 to 16 kD and for some reason our gels look very squiggly and ugly around that size.  Higher molecular weight bands look fine, the gel only begins to look that ugly around maybe 20 kD or so.  The dye front does not run very pretty though and begins to give that squiggly appearance when it is about halfway through the gel.  It has been very difficult to discern anything about our POI because of how ugly the gel is and I would really love any advice that could help me out here.  We're using a mini-protean tetra, and using pre-cast gels, 4-20%.  The gels are about a year old, but they aren't expired or anything.  We're using RIPA to lyse, a pretty standard recipe of it that I've found many places on the web.  After a BCA assay, I add Laemmli with 15% BME, boil at 100C for 10 min, load whatever amount I need to get 20 mg of protein from each sample onto the gel, and then run 120V for 1 hr.  We've also tried slowing it down to 60V for 2 hrs, with minimal difference.  We've made and remade the NaCl we're using in the RIPA to make sure its not a salt problem, and that hasn't made a difference either.  Any ideas here would be greatly appreciated, thanks!

#2 mdfenko

mdfenko

    an elder

  • Active Members
  • PipPipPipPipPipPipPipPipPipPip
  • 2,240 posts
77
Excellent

Posted 10 March 2011 - 12:10 PM

it's not so much the quality but the quantity of the salt that causes problems in page.

problems such as the ones you describe are usually caused by the buffer contributed by the sample. you can prove this by dialyzing away the ripa components (you can perform drop dialysis on a few microliters) and running that on the gel.
talent does what it can
genius does what it must
i do what i get paid to do

#3 JesseC

JesseC

    member

  • Active Members
  • Pip
  • 10 posts
0
Neutral

Posted 10 March 2011 - 04:50 PM

Thanks for the response mdfenko, I appreciate it.  We are using a 6x sample buffer, so I wonder if using a 2x buffer would dilute the salt out from the RIPA a bit more and help with this issue?

#4 mdfenko

mdfenko

    an elder

  • Active Members
  • PipPipPipPipPipPipPipPipPipPip
  • 2,240 posts
77
Excellent

Posted 15 March 2011 - 12:46 PM

it won't really help, you'll have the same mass of salt present. you can drop dialyze your samples to remove salt.