Hello, I'm new to this forum and to RT-time PCR.
I have been given the job to validate a real time PCR assay (4 singleplex reactions with sybr green).
The conditions of the reactions are per manufacturer's recommendations. I performed a standard curve analysis and here are the results:
1st primer set (amplicon size: 600 bp) - slope: -3.79 (CT for 1x template is 20)
2nd primer set (amplicon size: 570 bp) - slope: -3.48 (CT for 1x template is 26)
3rd primer set (amplicon size: 612 bp) - slope: -3.2 (CT for 1x template is 20)
4th primer set (amplicon size: 537 bp) - slope: -3.63 (CT for 1x template is 21)
My questions are as follows:
1) I'm going to keep the 3rd primer set. What about the other primer sets, should I try to optimize them or redesign them?
2) I read somewhere that if the slope is higher than -3.6 (ideal is -3.3), then I should look for other primer sets.
3) For primer set 2, the slope is within range, but the CT is not perfect, should I try to optimize this reaction, and if so, how (increasing Mg?)
I really appreciate your help. Any comments or suggestions are welcome. Thanks!
Clinical Real-time PCR assay
Started by tea, Dec 10 2010 09:18 PM
1 reply to this topic
#1
Posted 10 December 2010 - 09:18 PM
#2
Posted 30 December 2010 - 03:45 PM
The slope gives you an estimate of the PCR efficiency. Efficiencies over 100% (more than doubling every cycle) suggests contamination. Less than 100% suggests the reaction is working less well with each cycle. However, efficiencies estimated from the slope of a standard curve are also changed by inhibitors/enhancers in the standard curve dilution.
Efficiencies can also be determined at the individual reaction level using programs such as LinRegPCR.
Efficiencies can also be determined at the individual reaction level using programs such as LinRegPCR.