Weak western signal for help
#1
Posted 23 July 2010 - 01:43 AM
I need to undertake a lot of protein detections. At first, I use 4%-12% Nupage from Invitrogen, the band is great.
I detect some common protein in neuroscience study like PS-1 and APP series.
After about 3 times of using these commercial GEL, I began to cast sds-gel of Tris-glycine system my self.
I use similar electronic and antibody merit when I run Nupage and self-cast gel like Electrophresis:100v voltage consistant when I running 15% acrylamide/bis gel, the currency stays 22-26mA at stacking part and 12mA seperating part the process lasts about 140min. I always incubate my memebrane at 4 degree over night about 12 hours, and wash with 0.2% tween TBST before and after secondary antibody incubation 20minx3times.
The electronic bloting I use bio-rad minigel system with 55V voltage consistant and got 150-200mA in CAPS(pH11 with NaOH) bufer with 10% methanol for 85 min.
I carefully equilibrate the sds gel for 15 min before blotting also equilibrate membrane in methanol-MillQ-transfer buffer when I use PVDF. All my bands are developed with ECL reagent.
But the problem is weak signal, I tried 10%,12% and 15% gel several times, although I got better signal when I use gel with less density. But I need 15% to seperate a lot of proteins around 10KDa. When I use 15% gel to transfer the seeblue plus2 marker sometimes couldn't be fully transferred with miss of 250kDa and 148kDa. I use 100V consistant 120min for transfer once but the result is just a little bit better.
Thank you so much to read a problem so long. And I am waiting for your great opinions to help solve my problems.
#2
Posted 23 July 2010 - 04:05 AM
if transfer is proper, then check the antibody dilutions and incubation times... not forgetting wash systems and blockng!!!
Best luck!!
Cheers!!!
#3
Posted 23 July 2010 - 05:01 AM
I read a paper about reuse transfer buffer at least 7 times to maintain relatively good results(Transfer Buffer Containing Methanol Can Be Reused Multiple Times in Protein Electrotransfer Colin J. Pettegrew,* Renuka Jayini,* and M. Rafiq Islam). They even use the same buffer as SDS running buffer with 0.1% SDS.
I am totally confused. Sorry for so many questions.
#4
Posted 23 July 2010 - 11:31 AM
genius does what it must
i do what i get paid to do
#5
Posted 26 July 2010 - 05:42 PM
By the way, when I use ECL for detection, sometimes there was signal on film like flame, is it the resisted detection solution? How could I improve my operation? Thank you so much and sorry about so many questions.
#6
Posted 26 July 2010 - 08:28 PM
i was also wondering why wud u use a pH 11 buffer for transfer??!!! is the pI of ur protein too high say sumwere near 9 or so??!!! if not then the normal towbin buffer shud be fine.. cause may be at this high pH ur protein is getting degraded or sumthing!! (just a thought!!)
Edited by Prep!, 26 July 2010 - 08:32 PM.
Cheers!!!
#7
Posted 26 July 2010 - 10:38 PM
Prep!, on 26 July 2010 - 08:28 PM, said:
i was also wondering why wud u use a pH 11 buffer for transfer??!!! is the pI of ur protein too high say sumwere near 9 or so??!!! if not then the normal towbin buffer shud be fine.. cause may be at this high pH ur protein is getting degraded or sumthing!! (just a thought!!)
Some people use CAPS with PVDF for sequencing also CAPS only work between 9.5-11 pH. So I undertook this protocol. I use ECL on ice also I pre-open air conditioner before I use the film developer in dark room. It's wired because sometimes it occurs sometimes not.
#8
Posted 26 July 2010 - 10:45 PM
Cheers!!!
#9
Posted 26 July 2010 - 10:51 PM
Prep!, on 26 July 2010 - 10:45 PM, said:
Thank you for the rapid reply. I use wet transfer tank equipted with ice block. I will try Tris/Glycine transfer this time. Thank you
#10
Posted 27 July 2010 - 05:58 AM
what is the pore size of the pvdf and the size of the proteins of interest?
if your proteins are small then you may want to use a pore size of 0.22 um or smaller. if you are using 0.45 um then you may want to try a backup membrane to ensure that you are not blowing through (also, stain the gel to see if any is left behind).
genius does what it must
i do what i get paid to do
#11
Posted 29 July 2010 - 05:13 PM
mdfenko, on 27 July 2010 - 05:58 AM, said:
what is the pore size of the pvdf and the size of the proteins of interest?
if your proteins are small then you may want to use a pore size of 0.22 um or smaller. if you are using 0.45 um then you may want to try a backup membrane to ensure that you are not blowing through (also, stain the gel to see if any is left behind).
Thank you so much. I use 0.22um PVDF or nitrocellulose and my protein size ranged from 9kDa to 55kDa. The problem is when I stain the gel with commassie blue from Biorad, there are definitely bands on the gel after transfer.
The western is really puzzled, I tried same samples when I use NuPage from Invitrogen, the signal is more than 10 times different.
#12
Posted 30 July 2010 - 12:24 PM
mickey, on 29 July 2010 - 05:13 PM, said:
so you are not getting efficient transfer? are all the bands still in the gel (including the 9kDa)? (i would think that the 9 kDa protein would pass out of the gel rapidly and may pass through the membrane)
did you equilibrate the gel with transfer buffer before assembling the transfer sandwich (not always necessary but if you are not successful without it then you may want to try or vice-versa)?
did you ensure that there is full contact between the gel and membrane?
genius does what it must
i do what i get paid to do
#13
Posted 30 July 2010 - 05:37 PM
mdfenko, on 30 July 2010 - 12:24 PM, said:
mickey, on 29 July 2010 - 05:13 PM, said:
so you are not getting efficient transfer? are all the bands still in the gel (including the 9kDa)? (i would think that the 9 kDa protein would pass out of the gel rapidly and may pass through the membrane)
did you equilibrate the gel with transfer buffer before assembling the transfer sandwich (not always necessary but if you are not successful without it then you may want to try or vice-versa)?
did you ensure that there is full contact between the gel and membrane?
Thank you. Only bigger than 30kDa stained on the gel. I alwasys equilibrate the gel for 15 to 30 minutes. Also I cutted a 10ml pipet to get rid all the bubbles.
Now I could get good signal from 10% and 12% gel, also bands I want from 15% gel with little background. Problems are I now do some overexpression the signal seems not so good on 15% gel. Although other people's data are from NuPage or Tricine system, just because I want 9kDa and 70kDa on the same PAGE.
#14
Posted 02 August 2010 - 06:22 AM
mickey, on 30 July 2010 - 05:37 PM, said:
Now I could get good signal from 10% and 12% gel, also bands I want from 15% gel with little background. Problems are I now do some overexpression the signal seems not so good on 15% gel. Although other people's data are from NuPage or Tricine system, just because I want 9kDa and 70kDa on the same PAGE.
genius does what it must
i do what i get paid to do