very bad signals after sequencing


  • Please log in to reply
5 replies to this topic

#1 gyma

gyma

    Veteran

  • Active Members
  • PipPipPipPipPipPipPipPipPipPip
  • 146 posts
2
Neutral

Posted 07 June 2010 - 11:23 PM

dear everyone, I sequenced a plasmid DNA using AB 3130 sequencer and I got very bad signals and its not clear at all. I dont know what happened. can anybody give me some advice? my bigdye reaction is

plasmid: 2 ul (>200 ng)
bigdye (mixed with buffer): 2 ul
primer (1 mM): 4 ul
H2O: up to 10 ul

and PCR is

96C 1 min;
96C 10s,50C 5s,60C 4 min for 40 cycles.

and then I use ethanol/NaAc precipatation and resolve the DNA with formadine. and this protocol worked well before and I dont know why this time it didnot work.

#2 gyma

gyma

    Veteran

  • Active Members
  • PipPipPipPipPipPipPipPipPipPip
  • 146 posts
2
Neutral

Posted 07 June 2010 - 11:25 PM

sorry i forgot the attachment.

Attached Thumbnails

  • seq.jpg

Edited by gyma, 07 June 2010 - 11:27 PM.


#3 phage434

phage434

    Veteran

  • Global Moderators
  • PipPipPipPipPipPipPipPipPipPip
  • 2,747 posts
304
Excellent

Posted 12 June 2010 - 12:26 PM

It looks as if substantial amounts of labeled dNTPs are still present. I would be careful about the cleanup, perhaps using an additional 70% EtOH wash, or switch to a bead or column cleanup. You might also want to check the amount of template, which can be a problem with either too much or too little.

#4 mdfenko

mdfenko

    an elder emeritus

  • Active Members
  • PipPipPipPipPipPipPipPipPipPip
  • 3,335 posts
213
Excellent

Posted 18 June 2010 - 08:41 AM

how was the signal strength on the raw data?

if it was weak then you have a lot of noise.

if it was strong then you may be looking at multiple sequences.
if multiple sequence then you either have more than one primer, more than one priming site or more than one species of plasmid.

talent does what it can
genius does what it must
i used to do what i got paid to do


#5 gyma

gyma

    Veteran

  • Active Members
  • PipPipPipPipPipPipPipPipPipPip
  • 146 posts
2
Neutral

Posted 30 June 2010 - 07:50 PM

Thank you for your replies.
I put the same sample on another sequencer, and this signal pattern never appear again. although I still didnt get 100% clear result, the data is better than that from the previous sequencer. seems like the sequencer didnt work well.

#6 PandaCreamPuff

PandaCreamPuff

    Enthusiast

  • Active Members
  • PipPipPipPipPip
  • 43 posts
0
Neutral

Posted 04 August 2010 - 04:03 AM

does your sequencer get maintainanced according to the suggested timeframe for regular check ups? Our 3130 just had a maintenance done a month ago and had better results than ever before for everyone using it. Before it was alright but sometimes you would get very low signals and even failed sequence analysis. There are also the daily and weekly maintenance which should be done religiously to ensure quality results :)