bison, on Apr 13 2010, 05:41 AM, said:
Hi all,
I am trying to use the MMQPCR method of Cawthon to analyse telomere length but my LC480 is unable to cope with anlaysing the data which has takes two flouresence readings per cycle. I know that MyiQ analyser software will analyse the data but I don't have access to it, or the software for any other real-time analysers.
In essence, I have the floursence measured at each cycle for each of my products but I need software to adjust the baseline, fit a curve and calculate ct values.
Any/all suggestions about papers, online resources or offers to try my data on your software would be greatly appriciated and anyone that can help out would recieve appropriate acknowledgement/authourship.
Looking forward to hearing from you.
Hi Bison,
The following profile should work on the LC480:
1. 95 degrees x 15 minutes
-----------------------------------
2. 94 degrees x 15 seconds
3. 49 degrees x 60 seconds
Repeat steps 2 and 3, one more time
-----------------------------------
4. 94 degrees x 15 seconds
5. 59 degrees x 30 seconds
Repeat steps 4 and 5, 3 more times
----------------------------------
6. 85 degrees x 15 seconds
7. 59 degrees x 30 seconds, with signal acquisition
Repeat steps 6 and 7, 19 more times
----------------------------------
8. 94 degrees x 15 seconds
9. 84 degrees x 10 seconds
10. 85 degrees x 15 seconds, with signal acquisition
Repeat steps 8, 9, 10, 26 more times
----------------------------------
Melt program: 59 - 95 degrees; 0.5 degrees, 5 seconds, per step
Please note: during the cycling of steps 6 and 7, the telomere product is exponentially amplified, while the albumin gene product remains double-stranded and unavailable to its primers. Also, during the cycling of steps 8, 9, and 10 the albumin gene product is exponentially amplified, while the telomere product remains completely melted and unable to hybridize to its primers.
Primer sequences and concentrations:
telg and telc primers: sequences as in the attached paper, final concentration 700 nM each.
albumin primers (I recommend these over the albumin primers that are in the paper):
albugcr1: CggcggcgggcggcgcgggctgggcggAAAcGCTGCgCAGAATCCTTG, at final concentration of 200 nM;
albdgcr1: GcccggcccgccgcgcccgtcccgccgCTGAAAAGtAcGGTCGCCTG, at final concentration of 200 nM.
Serial dilution for standard curve: five concentrations, three-fold dilutions, with the highest concentration being 60 ng of DNA per 10 microliter reaction.
Master mix composition as in my 2009 NAR paper.
I observed no "No Template Control" amplification using the above protocol.
Richard Cawthon