has anyone tried to spin down nuclei and other cell debris AFTER adding laemmeli sample buffer (I forgot the spinning step - too many things going on)? I imagine SDS might crash out, but perhaps it could be added back. Thanks in advance!
spin hek cell lysate after adding laemmeli buffer?
Started by gellyfish, Mar 03 2010 04:31 PM
1 reply to this topic
#1
Posted 03 March 2010 - 04:31 PM
#2
Posted 05 March 2010 - 12:52 PM
the sds won't crash out. some of the debris will have solubilized so your sample will be more complex than it would have been if you had spun it first (it may also be more viscous due to the genomic dna from the nuclei).