Sandwich ELISA


  • Please log in to reply
1 reply to this topic

#1 shi

shi

    member

  • Members
  • Pip
  • 4 posts
0
Neutral

Posted 11 February 2010 - 02:14 AM

I am developing SANDWICH elisa to detect the amount of antigen in CSF samples. i coated different concentration of coating antibody and also prepared recombinnat standards. The result i get is as coating antibody concentration decreases the standard values decreases but CSF samples values increases as coating antibody concentration decreases.
The background is low.
Please let me know what is the problem.

#2 sgt4boston

sgt4boston

    Veteran

  • Active Members
  • PipPipPipPipPipPipPipPipPipPip
  • 276 posts
1
Neutral

Posted 12 February 2010 - 11:53 AM

What happens when you put your recombinate protein in "normal" CSF? Do you see the same response?


If your assay is working: antigen in buffer and in your matrix should produce parallel curves. If there is no matrix effect then the dose response curves will be on top of one another.


There are many many steps before you can go from just coating a plate with different ab coating concentrations to analysis of actual patient specimens.