hi, another question. Can anyone help me with this problem?
Does this have anything to do with the electrophoresis buffer?
the bromophenol blue band is skew during agarose electrophoresis
Started by fzhang, Oct 14 2009 03:45 AM
1 reply to this topic
#1
Posted 14 October 2009 - 03:45 AM
#2
Posted 14 October 2009 - 03:57 AM
This can happen if your buffer or gel is not level, or if the platinum electrode is broken. Anything that causes the current to flow preferentially on one side of the gel. Incidentally, I would switch from Bromphenol blue to Orange G as a dye -- it runs ahead of your DNA making the bands easier to see.