Help please kind experts.
We have an ELISA with the following protocol for a 30kD protein in unconcentrated urine.
Two monoclonal mouse antibodies were raised using the synthetically produced C-terminal 100 amino acids (Biosynthesis Inc, USA) as an antigen (Antibody Production Services Ltd, UK). One of these, APS1, was conjugated to Alkaline Phosphatase using the Lightning Link Alkaline Phosphatase Conjugation kit (Innova Biosciences, UK), whilst the other, APS2, was conjugated to biotin using the Lightning Link Biotin Conjugation kit (Innova Biosciences, UK). APS2-biotin was captured onto a 96-well streptavidin-coated plate (Nunc 436014, USA) at a concentration of 4μg /ml.
After washing, 100μl of urine or a dilution of the fragment in
buffer was incubated in each well for 1 hour at room temperature.
The plate was then washed 8 times in buffer and the secondary detection antibody – APS1-Alkaline phosphatase was added to each well at a concentration of 4μg/ml (1 hour at room temperature).
After a final wash step a colourmetric agent – pNPP (Sigma, USA) was added and the absorption of light at 405nm was measured after one hour. The dilution series was used to generate a standard curve by which the concentration of EN2 in each sample was measured.
To generate a decent looking standard curve we are having to make 5 measurements at each point and then take the median measurement.
We are also having to use 5 wells per sample and use a median to obtain a sensible output as individual wells vary by as much as a factor of 2.http://www.protocol-online.org/forums/style_emoticons/default/sad.gif
What are the most likely sources of our woes? All help gladly accepted
The protein we are measuring is sticky but that doesn't explain highish readings at 0 concentration.
Other ELISAs work fine on the equipment.