problem with transfer during western blot - western blot /sds page (Jul/01/2008 )
hi
 I am doing western blot and i am running  40ug of protein and run 100V for about 1.30 mins , i transfer in cold room at 100V  for 2 hrs. It worked fine for the first time but something is wrong with my transfer this time .
 my transfer buffer is 100ml of transfer buffer 1X, 200 ml methanol and 700 water.  i soak the gel and blot paper (PVDF) and whatnam ppaer from 5-10mins before asseble the wet transfer. When i check tranfer with ponceau i can see the top band  very faint  but not the small one, my protein is tranced protein so it ranges form 75 to 16 kd (each sample has different size) and i need to run in the same gel  please help.
Ps my gel is 12%
Hi
It all sound ok. Do u use the same membrane (pore size)?
U can also had SDS 2.5% or more to the transfer buffer.
The transfer buffer spoused to be X10 and not X1 as u wrote (am i correct?). cool the transfer buffer on ice 
before using, it helps.
try to load more protein on gel
good luck.
you should not use more than 0.05% sds in the transfer buffer. too much sds will interfere.